System not working on Medium and High Coag Setting
RE: Electrosurgical Unit (ESU) Report Post
Hello KenMed,
If it's a single use coag pen then try a new one.
Regards, Michel.
C. Michel Bessette, RRT (Canada #72-659)
Servicios Tecnicos CMB7
535 Canteval,
Ottawa, Canada
serviciostecnicoscmb7@gmail.com
Mobile: 6138185321
Bio-Medical Technological Services: Compressed Gas Systems Specialist
Anaesthesia & Hyperbaric Chamber Technologist
Services in: English, Français & Español
Bio-Medical Technological Services: Compressed Gas Systems Specialist Hyperbaric Chamber Technologist Services in: English, Français & Español
Report Post
Hello KenMed,
Your customer is not using a CUSA handpiece are they? If so they are activating the CEM circuit of the unit which will lower the power output. If not ask if the CEM light on the right side of the display panel is active. If it is, then there may be an issue with the CEM switch that is located behind the 4th pin on the first Monopolar handpiece receptacle.
Chris Jacobs
Product Development Engineer
800-221-9723 ext 124
Fax 818-882-0219
Replacement Parts Industries, Inc
625 Cochran Street
Simi Valley, CA 93065
Report Post
Hi Chris,
They are using this pencil, seems like a standard disposable. https://www.medline.com/sku/item/MDPESPB3002?skuIndex=S2&question=&flowType=&indexCount=
Thanks,
Evan
Report Post
Hello Evan & Chris,
Did that solve the issue?
Regards, Michel
Bio-Medical Technological Services: Compressed Gas Systems Specialist Hyperbaric Chamber Technologist Services in: English, Français & Español
Report Post
Hi all.
I have a force fx, its working ok in DESSICATE, but no output in FULGURATE or SPRAY. Cross checked control board with other working unit, all three coag settings working. Same control board showing this same behaviour in that faulty unit. So there is something wrong in motherboard (POWERSUPPLY/RF BOARD) of it.
Kindly guide me so that i resolve this issue.
Thanks.