shock equip malfunction

I keep failing the operational test with fail/pads- There is an error code of '' shock euip malfunction'' when unit is turned on- I have replaced the Therapy PCA as suggested in the service manual- new therapy cable as well and new test load cell. Any ideas on what is going on with this unit? Thanks in advance-

   RE: Defibrillator Report Post
Tue Sep 22 2020Reply from ReNew14
avatar placemark

It's probably the capacitor. 


Report Post
Wed Sep 23 2020Reply from DFMK6
avatar placemark

It sure was the capacitor.

 

Thanks ReNew14

 


Report Post
Wed Sep 23 2020Reply from ReNew14
avatar placemark

Your very welcome. 

if you ever need defibrilator parts for the MRX, zoll, physio control ,

contact ReNew Biomedical 844-425-0987 

 