error code 9022
RE: Mobile C-Arm Report Post
Error W9022 Radiation release conditions not allowed
Possible trouble: Release conditions for radiation release are not fulfilled.
Error detected during boot up.
Possible help: If the error W9022 is stored without additional error messages:
User fault: Footswitch released during boot up.
Maybe foot or hand switch defective.
Check hand switch (D3 / D30, LED V12)
Check cables of the hand switch for short circuit or wire breakage.
Check foot switch (Left pedal D3 / D30, LED V10 and right pedal D3 / D30, LED V30)
Check cables of the foot switch for short circuit or wire breakage.
If additional error messages are stored (Error messages W9010 to W9024),
search and read the error message descriptions of the additional error message.
-----------
Error E4720 Memoskop no response from 3D PC
Possible trouble: The Memoskop registered an communication error with the 3D PC
Possible help: Is the 3D PC running?
Check the 3D PC and restart the system.
Check connectors Memoskop - 3D PC
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Hello Wayne thank you for a super fast reply, I will check the led's on V12 , 10 and 30 and report back to you. Many thanks because it is impossible to get a tech to come down because of covid 19 restrictions here in Antigua. Once again thank you and have a great evening.
Pat
Report Post
Fixed. it was blown fuse controling hand and foot switch thanks for the advice worked great
Report Post
You are very welcome. Glad things worked out. - Wayne