error code 9022

Greetings all. today i got an error code 9022 followed by code4720 a few minutes later if the unit stays on. The unit starts up but will not allow normal function when we used it yesterday everything was normal and functional. what does code 9022 and 4720mean and what can be done to solve it. your guidance in this matter is deeply appreciated.

   RE: Mobile C-Arm Report Post
Fri Sep 25 2020Reply from outinleft
avatar placemark

Error W9022 Radiation release conditions not allowed

Possible trouble: Release conditions for radiation release are not fulfilled.

Error detected during boot up.

Possible help: If the error W9022 is stored without additional error messages:
User fault: Footswitch released during boot up. 

Maybe foot or hand switch defective.

Check hand switch (D3 / D30, LED V12)

Check cables of the hand switch for short circuit or wire breakage.

Check foot switch (Left pedal D3 / D30, LED V10 and right pedal D3 / D30, LED V30)

Check cables of the foot switch for short circuit or wire breakage.

If additional error messages are stored (Error messages W9010 to W9024),

search and read the error message descriptions of the additional error message.

-----------

Error E4720 Memoskop no response from 3D PC

Possible trouble: The Memoskop registered an communication error with the 3D PC

Possible help: Is the 3D PC running? 

Check the 3D PC and restart the system.

Check connectors Memoskop - 3D PC


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Sat Sep 26 2020Reply from nsa
avatar placemark

Hello Wayne thank you for a super fast reply, I will check the led's on  V12 , 10 and 30 and report back to you. Many thanks because it is impossible to get a tech to come down because of covid 19 restrictions here in Antigua. Once again thank you and have a great evening.

Pat


Report Post
Tue Sep 29 2020Reply from nsa
avatar placemark

Fixed. it was blown fuse controling hand and foot switch thanks for the advice worked great


Report Post
Tue Sep 29 2020Reply from outinleft
avatar placemark

You are very welcome. Glad things worked out. - Wayne