Error 9022

We have a C-Arm and have error 9022

   RE: Mobile C-Arm Report Post
Wed Oct 07 2020Reply from outinleft
avatar placemark

Check your handswitch, footswitch, and D3. - Wayne Horsman

------

Error E9022 Radiation release conditions not allowed

Possible trouble: Release conditions for radiation release are not fulfilled.

Error detected during boot up.

 

Possible help: If the error W9022 is stored without additional error messages:
User fault: Footswitch released during boot up. 

Maybe foot or hand switch defective.

Check hand switch (D3 / D30, LED V12)

Check cables of the hand switch for short circuit or wire breakage.

Check foot switch (Left pedal D3 / D30, LED V10 and right pedal D3 / D30, LED V30)

Check cables of the foot switch for short circuit or wire breakage.

If additional error messages are stored (Error messages W9010 to W9024),

search and read the error message descriptions of the additional error message.


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com