Lumenis Selecta Duet

E122 Energy Low Fail. Which card could be the problem?

   RE: Surgical Laser Report Post
Wed Oct 07 2020Reply from Biyomedikalege
avatar placemark

E122 Energy Low Fail. Which card could be the problem?


Report Post
Wed Oct 07 2020Reply from Laserdude
avatar placemark

E122; Energy low Fail

Energy of the last pulse fired was less than the expected energy by more than 70%.

Requires recalibration of energy levels.

boards are not suspect here

 


Report Post
Mon Oct 12 2020Reply from Biyomedikalege
avatar placemark

 

Hi;

What is the necessary procedure for calibrating energy levels?

I would be glad if you help for this.

If there is a service manual, can you send it?

Best Regards.

 

ZEYNEDDIN KARAASLAN

Biyomedikal Technician

 


Report Post
Mon Oct 12 2020Reply from Ben C
Ben C

Have you checked the MANUALS


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Mon Oct 12 2020Reply from Laserdude
avatar placemark

You need a Coherent LabMax Top Power/energy meter.  A 20 joule pyroelectric energy detector for it as well as the cable to attach to the instrument and the manual to follow the procedure.


Report Post
Mon Oct 12 2020Reply from Laserdude
avatar placemark

 Without the proper equipment and experience in using the equipment any attempt at calibration would probably be detrimental to the instrument.