Lumenis Selecta Duet
Wed Oct 07 2020Reply from Biyomedikalege
Wed Oct 07 2020Reply from Laserdude
E122; Energy low Fail
Energy of the last pulse fired was less than the expected energy by more than 70%.
Requires recalibration of energy levels.
boards are not suspect here
Report Post
Mon Oct 12 2020Reply from Biyomedikalege
Hi;
What is the necessary procedure for calibrating energy levels?
I would be glad if you help for this.
If there is a service manual, can you send it?
Best Regards.
ZEYNEDDIN KARAASLAN
Biyomedikal Technician
Report Post
Mon Oct 12 2020Reply from Laserdude
You need a Coherent LabMax Top Power/energy meter. A 20 joule pyroelectric energy detector for it as well as the cable to attach to the instrument and the manual to follow the procedure.
Report Post
Mon Oct 12 2020Reply from Laserdude
Without the proper equipment and experience in using the equipment any attempt at calibration would probably be detrimental to the instrument.