Mon Oct 12 2020

Booting error 1-0-0

I kept getting Booting error 1-0-0 cant get to the scanning window. Desperate in the jungle can somebody help me? Do I need to insert the MO disc to get to the scanning window.?

   RE: Ultrasound Systems Report Post
Tue Oct 13 2020Reply from dferguson
dferguson

1 - 0 - 0 points towards a power issue.  Likely this is a MDI issue but could also be the PPS or Fan Tray.  Measure your power at the outlets inside the rear access panel.  If getting 120VDC, then Fan tray should be ruled out. Measure on the right side of the system out of the MDI Power supply and verify there is 120VDC. Measure your DC Voltages and see if any are missing. 


 
 

Report Post
Wed Nov 24 2021Reply from haden
avatar placemark

Hi Mr, Devid,
I have problem with Ecoh probe 5V2c when i contact in the machine give me inveled how i can repair it,


Report Post