Blowing a fuse in F1 or F2 fuses. Battery does not seem to have any charge on it. Does anyone have any ideas? The fuse was a 250v 6amp time delay fuse, the one that keeps blowing on the right side, once fuse was back in holder
BLOWING FUSE IN F1
What is the name of the manufacturer and model of the equipment you are working on?
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
you've got a short somewhere, or your plugged into the wrong outlet.
check the current with the table on battery only
verify the voltage at the wall
check your power supply, battery contacts, etc.
Report Post
Isolate the A/C plate assembly (power entry w/fuses) by unplugging P17 at the power supply. Remove shrouds and follow the brn/org wires to the power supply, P17 should be near the large blue capacitor mounted on the side of the power supply. If the fuse still blows replace the A/C plate, if it only blows when power supply is plugged in, the power supply is faulty. Replacing the power control board in the power supply could solve the problem but I normally just replace the whole assembly.