Air detected in line

I have two used Baxter Sigma Spectrum infusion pumps that were sent out for service. The technician reported that each is reporting the same error message: not recognizing that the tube is in place, air detected in line. He stated that the sensors have been cleaned. He contacted Baxter and was informed that that each requires a new board ( that will exceed the value of the units!) Any ideas what may be causing this on two units?

   RE: Infusion Pump Report Post
Wed Oct 21 2020Reply from Chris S
Chris S

There are multiple 3rd party companies that are able to repair these units at a reduced cost.  The company I use typically replaces the oscillators when this problem is noted.  I have not had any luck locating this (or most)  parts for purchase through Baxter or any 3rd party service.  I would be curious if anyone else on here has.


Official MedWrench Guru
Report Post
Wed Oct 21 2020Reply from Jurmedica
Jurmedica

Chris - thanks for the info and insight. Can you advise contact info for any of these companies?

Thanks

Harold  - jurmedica@gmail.com


H

Report Post
Wed Oct 21 2020Reply from Ben C
Ben C

MedWrench works as the link between you and a service company. Browse service companies for your equipment or fill out a REQUEST FORM and we will help find a service company for you! 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Fri Oct 29 2021Reply from JPJ
avatar placemark

I've received this error before when testing the device if you're re-using IV lines during testing. If you're testing with an older IV line, toss it and use a new one for testing. 

If an end user is receiving this error with a new IV line, then the load point 2 sensor will need to be cleaned gently (no sharp objects). After it's been cleaned, I would suggest calibrating the sensor as well. 

A 3rd party Sigma Infusion pump repair company I love to use is Elite Biomedical. They always resolve the issue with a quick turn around time. Hope this helps!


Report Post
Tue Dec 06 2022Reply from salarypouch
avatar placemark

If you are reusing IV lines while testing the gadget, I have previously had this issue. Use a fresh IV line for testing instead of an old one if you are using one.


run 3

Report Post