Time and Date issue

The coin cell is at 3.25 VDC. The hardware test states that system memory and static memory both errored out. Is this a coin cell issue or the main board? Any assistance would be greatly appreciated.

   RE: Electrocardiograph (EKG / ECG) Report Post
Thu Oct 22 2020Reply from Ben C
Ben C

One of these companies may be able to assist:

Southeastern Biomedical

GE


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Fri Oct 23 2020Reply from S. Rooks
avatar placemark

Hello Ahamayym,

Among other things, the coin cell battery maintains the time/date and changes to default settings. If the cell voltage maintains @ 3.25, is it possible you have a cold solder joint? To troubleshoot, I would suggest checking the voltage along the traces leading to the battery posts. Often I have encountered boards with -/+ solder points that have been torn or damaged when individuals attempted to replace the coin cell battery.


Biomedical Equipment Technician with Southeastern Biomedical 1-888-310-7322
Report Post
Fri Oct 23 2020Reply from Ahamayym
avatar placemark

I verified the battery voltage at the battery and through the board at the solder points.  Unfortunately, it isn't a cold solder joint.  Thank you for the recommendation.  I am thinking that the main board has finally gone bad.  It has a mfr date of 2008.