am kindly requesting for the product manual
Wed Oct 28 2020
manual
Wed Oct 28 2020Reply from Ben C
MedWrench is available to help people with specific questions and troubleshooting problems. Could you please be more specific with the issue that you are having with your instrument, and the MedWrench team will do their best to assist you in solving the issue. This will also help other community members who face similar challenges.
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
Thu Oct 29 2020Reply from Kennerdol
When I run the Control the PT is out or range (I ran 4 tests and the results were 1. 29.0sec, 2. 28.3sec, 3. 28.6sec, and 4. 28.6sec)
and running the Sample(2 tests) the results were 1. 0.8sec and 2. 7.4sec
The range we use for PT:
- UNITS: Sec
- MEAN: 11.1
- RANGE LOW: 8.88
- RANGE HIGH: 13.32
NOTE: I changed the Control Solution about 4 times.
Report Post