I keep getting "Check Pads" every few seconds. When I set up the 30J test, the unit will not discharge, charges, but wont discharge. Forget about higher settings, unit will not discharge at any setting. I tried another multi-function therapy cable, but no luck. Anyone seen this before? Any suggestions?
Wed Oct 28 2020
Check Pads
Wed Oct 28 2020Reply from ReNew14
I've seen it be the right side panel/ test port
seen the PD Engine cause this.
Possilbe even the anlalog system pcb.
We can help you with this at ReNew Biomedical 844-425-0987
obviously the test port is the easiest. do you have an old sytle test plug from the m or e series.? you can use one of those to verify.