Force 40 unit. Several of my resistors are incorrect compared to the manual's parts list.
R97,98 150 ohm, 50W according to the manual, unit contains 150 ohm and 100 ohm 50w resistors. The manufacture of both resistors is the same suggesting factory OEM.
R14,84 100 ohm 1/4 watt according to the manual, when the unit contains a 56 ohm resistor much bigger than 1/4 watt in the r84 slot. R14 is correct according to the manual at 100ohm 1/4 watt. The PCB painted outline for 84 is much larger than a 1/4 watt resistor, suggesting this is the factory OEM resistor.
Finally R69,70 are 510 ohm 1/2 watts according to the manual but 69 is a 4.6k ohm possibly 2watt? based on size. 70 was burnt and I cannot confirm size but the outline confirms it should be much bigger in wattage than the manuals stated 1/2 watt.
Attempting to replace R70 with a 4.7k ohm 2 watt does nothing advantageous. I am getting codes
#3
#86
#100-104
#111-118
#122
#124
Any help on the color code of R70 would be helpful. 4.6k ohm 2 watt resistors look to be unobtainable on the resistor market.