What does error 42 mean in Zeiss Visulas 532s
Wed Nov 04 2020
Error 42
Fri Nov 06 2020Reply from Wayne-UML
It is a software comparison fault. Parameters are compared to a stored value, which in your case is not matching. You should look at replacing the NVRAM.
Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Mon Nov 09 2020Reply from LeonidICEM
this error (42) is ready resolved, thanks so much for your reply, it was very helpful. Now I have two new errors, 86 and 96 and I don´t know what they mean, I´m looking for the service manual of the VISULAS 532s system, is there any chance you could help me to get it?
Tue Nov 10 2020Reply from Wayne-UML
I have sent the manual to both Kari and Mandy. I believe one of them will be uploading it onto the site.
Wayne Delgado
www.umlasers.com
Wed Jun 30 2021Reply from xicayuda
BUEN DIA.
MI EQUIPO ME ARROJA EL ERROR 40 AL ENCENDERLO ¿ALGUIEN SABE QUE LE PUEDO HACER?