Hi, I got problem with Siremobil Compact L that show white screen in image area any KV with no error code sometimes. I changed Camera already but problem still. Please advice. Regards, Chao.
P.S. the camera I swapped is in stock that not newly one.
White display with no error code.
HI - Is this Chaovalit Jitsinthu at MicroMed in Thailand?
Is the white screen in both displays? if only one, which display?
when the white screen is showing, does the kv still regulate (with changes in patient density)?
- Wayne Horsman
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Hi Wayne Horsman,
I'm chaovalit from Thailand. How are you.
The white screen happen sometime. When happen the X-ray can generate any KV.
I guest to problem come from Memoskop So customer use new desk fan to cool down the Memoskop and the problem not yet comes (about 1 week ago every day use).
Hope you are healthy and happiness.
Regards,
Chao.
Hi, do you have the model and brand of this camera ?
where did you found used cameras for siremobil compact L c arm
thanks a lot
Hi, juniorlaxe,
Sorry for reply late. I just came from long vocation.
Used camera I got from another defective unit.
Regards,
Chao.
It is easy enough to isolate the camera from the image intensifier and the Memoskop by testing them *while the symptom is occurring* You can point the camera at the wall & see if you get an image. you can look at the output phosphor of the image tube (during exposure) to see if the is an image there (is phosphor lighting up?), etc.
Wayne Horsman
Columbia Imaging