Mindray BC 2800

What could be the problem when we have RBC and WBC bubbles?

   RE: Hematology Analyzer Report Post
Sun Nov 29 2020Reply from Engr Michael Okeke
avatar placemark

Please check that your Lyse is in good condition (Not expired, Not empty, Properly aspirated)


Report Post
Fri Jan 22 2021Reply from GeorgeM
avatar placemark

It might be a problem with the angled tubes,at the bath. From time to time is a good ideea to replace the small 4 tubes  that are bringing Rinse in the circuit. 
Take out the black metal box, the bath and you will have acces to it. 
When these tubes are not perfectlyy fitted on the 2 white elbows and the T piece is a great chance of bringing air into circuit. 

 


Report Post