error 4017
- what is your email address?
- Usually a bad Memoskop (if it was working normally prior to failing)
Error E4017 Memoskop busy signal timeout
Possible trouble:
Host registered timeout of the busy telegram task,
Possible help:
Sporadic: Faults in the system, Memoskop timing too slow.
Continuous: D1 - Memoskop software version not compatible, Memoskop defective.
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
E4017 with radiation release.
Without the generator conencted does all the Memskop functions work, like ctr + U & ctr + T?
Does the text appear on screen?
Does the memoskop function (as above) work with generator connected?
Check the X10 'main umbilitcal cable' connection pins for damage/bent over, make sure none of the pins are pushed-in, a couple of pins are partly-in, this is normal.
Check the cable wires between the X10 & Memoskop in case a wire is broken.
Tom
Hi, is there anyone here could help me where to buy Raycam for Siremobil Compact L part no. 08887452 or 08888021.Thanks