Error 4731 4028 5032
RE: Mobile C-Arm Report Post
Dan - send me your email & I will see if I can help you. wayne@columbiaimaging.com
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
HI
Error 4731: This is missing Dongle.
When you replaced the Memoskop did you also move the Dongle from faulty Memoskop to new Memoskop?
Is the new Memoskop & old Memoskop using the same software? Did you load the software onto the new Memoskop?
The new & old Memoskops, are they the same part number, there are two part numbers, check both.
Error 4021: Mean the Memoskop did not answer!
Error 4028: Unexpected Acq end signal, possible because of E5032.
Just as a quick test, put some Cu on tube (2.1mm Cu) & see if you get error 5032? If no it could be generator learning/kV offset adjustment.
What was the original reason Memoskop replaced?
As always, check the X10 pins can get bent or pushed back, a cople of pins are slightly reseassed, which is normal!
Tom
Report Post