There is no image from the endoscope camera on the display.

Hello. Video processor Fujinon VP-4400. There is no image from the endoscope camera on the display. Where the image should be, a dark background with noise. The endoscope is serviceable, checked on another video processor. Help if you know what is the cause of the malfunction. Where can I find the service manual? Thank you.

Report Post
Tue Jan 12 2021Reply from Chris-CBET
avatar placemark

is your light source working?


Report Post
Wed Jan 13 2021Reply from Vladimir.s
avatar placemark

Yes, light source working.


Report Post
Wed Jan 13 2021Reply from Vladimir.s
avatar placemark

Yes, light source working.


Report Post
Wed Jan 13 2021Reply from Chris-CBET
avatar placemark

You said the scope is serviceable? You tried it on a different machine? Are you saying you have verified the scope is not the issue?

Have you tried using a diff scope on this unit?

Check the connection point where the scope plugs in. 

Check connections in your camera unit. 


Report Post