Hello. Video processor Fujinon VP-4400. There is no image from the endoscope camera on the display. Where the image should be, a dark background with noise. The endoscope is serviceable, checked on another video processor. Help if you know what is the cause of the malfunction. Where can I find the service manual?
Thank you.
Tue Jan 12 2021
There is no image from the endoscope camera on the display.
Tue Jan 12 2021Reply from Chris-CBET
Wed Jan 13 2021Reply from Vladimir.s
Wed Jan 13 2021Reply from Vladimir.s
Wed Jan 13 2021Reply from Chris-CBET
You said the scope is serviceable? You tried it on a different machine? Are you saying you have verified the scope is not the issue?
Have you tried using a diff scope on this unit?
Check the connection point where the scope plugs in.
Check connections in your camera unit.