Fri Jan 29 2021
Repair
Unit will not go up or down when up and down buttons are pushed. Unit tilts forward instead. All other functions are operational.
   RE: Bed Report Post
Fri Jan 29 2021Reply from Amscopolaris
Amscopolaris
Trendelenburg Switch Active, Turn it off Trendelenburg Switch
Fri Jan 29 2021Reply from Chris-CBET
Chris-CBET
Either your rev trend is engaged, or you have a bad actuator. Check to see if you have voltage to the foot end actuator. See if the motor is  spinning, if it is spinning, you have a bad coupling
Fri Jan 29 2021Reply from ADJBED-MD
You are correct. The coupling is damaged.  Thank you all for your assistance.