Tue Feb 02 2021

Gas analyzer_ AION Platinum

Due to not emptying the water trap, in the Maquet Flow-i C20, water return back to gas analyzer module. Is there a way to repair this module or needs to be replaced. Please advise.

   RE: Anesthesia Gas Machine Report Post
Wed Feb 03 2021Reply from Clint Ech
avatar placemark

The gas analyser module can be stripped down to component parts. You may be able to save it by doing this. Worst case is that you can't fix it and would have to buy an expensive replacement. we have had success with repairing them but we haven't had this particular issue.


Report Post
Wed Feb 03 2021Reply from ialmajali
avatar placemark

Thank you Clint.

If worst case happened, do you have an idea source of suppliying this parts chaeper than Manufacturer.

Regards


Report Post
Wed Feb 03 2021Reply from Clint Ech
avatar placemark

We have spoken to the manufacturer of the gas analyser and they will not sell the units to us directly, they will only supply them to Maquet (Getinge) so we had to go the Maquet for a replacement when we had one that we couldn't repair.


Report Post