Error 37 measn no kV or short on the kV inverter.
There is the main fuse 80 or 100 amps which blows & give this error, its fairly common.
The fuse is mounted on main inverter assembly, it can be located in two different positions depening upon the board version.
Lucky version/location: The fuse can be accessed from the side, remove side covers (left side) & the fuse could be sisible through hole.
Be very careful as the capacitr bank could remain charged, measure the voltage on both sides of fuse first.
Unlucky version/location: The fuse is located in the middle of the capacitor board, you have to remove all plugs & wires to inverter unit & slide-out to access fuse & boards.
There is also a 20amp fuse from cap charger that could give same syptoms!
The servcie manuals can be downloaded from the Siemens Healthineers whebsite after registering.