Error message: Err 37

Hi, the display is showing me Err 37, I would like to know why this error is due. And if anyone has the service manual for this equipment that you can send me. My email is juniorlevanogonzales@gmail.com

   RE: DR Systems Report Post
Wed Feb 10 2021Reply from Tomelgy
avatar placemark

Error 37 measn no kV or short on the kV inverter.

There is the main fuse 80 or 100 amps which blows & give this error, its fairly common.

The fuse is mounted on main inverter assembly, it can be located in two different positions depening upon the board version.

Lucky version/location: The fuse can be accessed from the side, remove side covers (left side) & the fuse could be sisible through hole.

Be very careful as the capacitr bank could remain charged, measure the voltage on both sides of fuse first.

 

Unlucky version/location: The fuse is located in the middle of the capacitor board, you have to remove all plugs & wires to inverter unit & slide-out to access fuse & boards.

There is also a 20amp fuse from cap charger that could give same syptoms!

The servcie manuals can be downloaded from the Siemens Healthineers whebsite after registering.

Tom