Sun Feb 14 2021
NBP pump malfunction
Originally the user complained that the diastolic pressure reading was too high. Attempted an NBP calibration but the pump did not come on. I've replaced the following: - NBP pump module - Mainboard - Power supply module I am still getting an NBP malf and pump is not working which is causing it not to calibrate. https://omegle.onl/
   RE: Patient Monitoring Report Post
Sun Feb 14 2021Reply from Chris-CBET
Chris-CBET
What is the make model?
Sun Feb 14 2021Reply from Ali Youssef
Ali Youssef
Can you let me know the model number of the unit please? We will try to help.
thanks
USOC Medical
Mon Feb 15 2021Reply from T-Lobato
T-Lobato
Perhaps the NBP pump module was bad.  Are you able to take voltage readings (at the pump) when it should be engergized?
TJ Lobato | MedWrench Guru