Sun Feb 14 2021

err 4 (NF 400)

according to the manual "Err 4: Motor overheat failure. Please wait for the motor to cool down and start the centrifuge again. " We waited for 3 hr now and the error still appears, what we should do?

   RE: Centrifuge Report Post
Tue Feb 16 2021Reply from Amscopolaris
Amscopolaris

1 Clean the ventilation channels, ( Generally, as there is dust on the cooling condenser, cooling will be late.)

2 PT 100 Check the temperature sensor 

3 Replace   PT 100


Report Post
Tue Feb 16 2021Reply from Ben C
Ben C

Ozark Biomedical may be able to assist. 