NIBP Leaking

NIBP Leaking not out side unit, any ideas on what could be the cause.

   RE: Vital Signs/Multi-Parameter Report Post
Wed Feb 17 2021Reply from T-Lobato
T-Lobato

Perhaps one of the hoses disconnected inside the unit.

T.


TJ Lobato | MedWrench Guru
Report Post
Wed Feb 17 2021Reply from Hill452
avatar placemark

It is just a slight leak, not like a hose is off, It will take BPs but fails the leak test.

Thanks,

Brian


Report Post
Wed Feb 17 2021Reply from Ben C
Ben C

One of these companies may be able to assist:

USOC 

Southeastern 

Southwestern

Sage Services 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Wed Feb 17 2021Reply from Chris S
Chris S

The Accutor Plus had 2 valves that were notorious for creating small leaks.  They had to be taken apart and cleaned on a regular basis.  I'm not sure if the same applies to the V,