Errors displayed
RE: Blood Pressure Monitor Report Post
Is the device cycling for an NIBP (Non-Invasive Blood Pressure) measurement at all, or do you just immediately get an error and cannot proceed? The e-20 indicates your NIBP is out of calibration. Have you tried different hoses and or/cuffs? You could have an internal hose disconnected or a compressor issue. I would give Smith's a call. They're helpful!
Deanna Wilkinson - MedWrench Guru
Report Post
Thanks for the response. On power up I get E-4 consistently, then when measurement button pushed, within 1 sec I get E-20 with no sound from the pump.
I dong get a very good response from Smiths. Most of the time it is : send it in with 4-8 weeks turn around. I was trying to get CO2 modules for Advisor monitor, and after 2 weeks trying to get hold of a sales guy to authorize it, I was told its not for sale, only for internal use.
Thanks again
Vladimir
Report Post
Thank you so much by starting this thread you have solved many others people issue which are using the same version of it.