Tue Feb 23 2021

Trouble Shooting

has anyone run into an error code of iiiii iiii Turn on the pump and I get a bunch of iiii iiii iiii

Report Post
Wed Feb 24 2021Reply from Ben C
Ben C

One of these companies may be able to assist you:

Elite Biomedical 

Mobile Medical Repair 


Find me on FacebookLinkedIn, YouTube or Twitter 
Email me: info@medwrench.com

Report Post
Wed Feb 24 2021Reply from garbleduser
avatar placemark

If memory serves, this error is caused by a dead internal lithium memory backup battery soldered to the mainboard. It is depicted on page 79 of the service manual. It is on the extreme lower left in the board silk screen layout graphic. Just make sure to get the polarity correct when resoldering the new battery in place. Link to service manual: 

https://cdn.medwrench.com/eMan/equipmentNo/8842/sigma-8000-plus-service-manual.pdf

I hope this helps!

-Garbleduser


Report Post