Wed Mar 03 2021

Oxygen Percentage only 30%

I have an Everflo that the oxygen percentage only gets up to 30%. The Sieve canisters were replaced about 2000 hours ago so I assumed it was the compressor. However when I replaced the compressor there is no change. There does not appear to be any leaks (I fixed the ones I found). Whats next?

   RE: Oxygen Concentrator Report Post
Mon Mar 08 2021Reply from Chris-CBET
Chris-CBET

how are you verifying the O2 output? 


Report Post
Mon Mar 08 2021Reply from CarlAnderson
avatar placemark

I use a Salter Oxygen Concentrator analyzer. I have calibrated it and it does work fine on other concentrators.

Salter Elite 2 Concentrator Indicator


Report Post
Mon Mar 08 2021Reply from Chris-CBET
Chris-CBET

is the unit alarming low O2 or is this purely based upon your analyzer?


Report Post
Mon Mar 08 2021Reply from Chris-CBET
Chris-CBET

if the unit isn't alarming, but your tester says it's no good. I'd question the validity of your tester. 

if it is alarming, then I would suggest checking the intake filter, the main valve, maybe it's not switching between sieve beds or allowing enough intake. 

after that, the valve b/w the sieve beds and the product tank, the product tank itself, anyone of the check valves, and the bacteria filter


Report Post
Mon Mar 08 2021Reply from CarlAnderson
avatar placemark

It does alarm after a few minutes of operation. Pressure is cycling, no change with changing filters. I suspect its the sieve beds themselves, maybe moisture contaminated. I have ordered a new sieve canister assembly


Report Post
Mon Mar 08 2021Reply from CarlAnderson
avatar placemark

It does alarm after a few minutes of operation. Pressure is cycling, no change with changing filters. I suspect its the sieve beds themselves, maybe moisture contaminated. I have ordered a new sieve canister assembly


Report Post
Mon Mar 08 2021Reply from CarlAnderson
avatar placemark

It does alarm after a few minutes of operation. Pressure is cycling, no change with changing filters. I suspect its the sieve beds themselves, maybe moisture contaminated. I have ordered a new sieve canister assembly


Report Post
Thu Mar 11 2021Reply from Ameer alkadhimi
avatar placemark

Do you solve this problem?

I think is O2 sensor if you give low o2


Report Post
Fri Mar 12 2021Reply from Chris-CBET
Chris-CBET

he's getting the O2 alarm on the unit and his external testing device is confirming the low O2 output. it' snot a snesor issue unless both sensors are bad. It's more likely a leak or closed valve issue


Report Post
Wed May 05 2021Reply from CarlAnderson
avatar placemark

Sigh so i think now that it is the control board, as really its the only thing that hasn't been replaced - new compressor, new sieves, new valves and no change.


Report Post