how do i solve this error on cr 10x the ready light is always red

   RE: Computed Radiography Report Post
Thu Mar 04 2021Reply from Luis_Med
avatar placemark

Hello,

I hope all is well. Did you tryto reboot the CR?. Are you getting error messages?.

Regards,

Luis 


Report Post
Thu Mar 04 2021Reply from Luis_Med
avatar placemark

I just saw the error. You need to do the follow:

Restart the Digitizer and check

2. If error still persists replace optic module";"OM_OperatingConditionErr";"ARC2101";"Optic Module operating conditions are not within specification.

Meaning: The 'Optic Module' setting and adjustment are out of specification.
Or the Galvo parameters are not adjustable.
Or the Laser is not working stable.
Or the detection of begin of scan failed.
Or an optical line synchronisation signal is missing during scan.

Reason 1: The 'Optic Module' synchronisation failed because of mechanical shock.
Cure 1: Restart Digitizer to get detailed selftest result.
Try to scan an IP again to test the 'Optic Module'.

Reason 2: The software has no current settings for the 'Optic Module'.
Cure 2: Install the newest digitizer software.
Restart Digitizer to get relevant selftest result.
Try to scan an IP again to test the 'Optic Module'.

Reason 3: The cable to 'Optic Module' is defective or disconnected.
Cure 3: Check cable and connectors between 'PMI Board' and 'Optic Module'.
Replace 'Cable PMI Board to Optic Module'.

Reason 4: The 'Optic Module' is defective.
Cure 4: Replace 'Optic Module'.

Reason 5: 'PMI Board' defective.
Cure 5: Replace 'PMI Boar

 