no longer heating
RE: Hot/Cold Therapy Systems Report Post
What is the name of the manufacturer and model of the equipment you are working on?
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
I am sorry. I had put that info in the title but it didn't transfer.
This is a Gaymar T pump 500 model. It still pumps fine and holds water. The water, however, is not hot.
Report Post
It could be the heater or the PC board. Check if you have 120 Volts going to the heater (red and blue wires). If you have voltage, it's the heater. No voltage, it's the board. Also make sure the overtemp light is not on and the unit is on a level surface.
Also make sure it is pumping water. Alot of complaints I get about "not heating" are because it is actually not pumping any water.
Official MedWrench Guru
Report Post
Thank you. That is just the info I needed. When I figure out how to use my voltmeter,(approx from the 1950s), I will proceed .
Madelyn C
Report Post
Hmmm, I want to be sure that I have this right. I get NO voltage across the heater. Wouldn't that indicate that the heater needs to be replaced?
Report Post
The PC board supplies the power to the heater, so if there is no power it is the PC board.
You can also take a resistance measurement across the heater with the heater disconnected. It should read approximately 80 ohms. That would help confirm it.
Official MedWrench Guru
Report Post
The PC board supplies the power to the heater, so if there is no power it is the PC board.
You can also take a resistance measurement across the heater with the heater disconnected. It should read approximately 80 ohms. That would help confirm it.
Official MedWrench Guru
Report Post
Yes, I confirmed it. The heater has plenty of life. Now, to find a circuit board! You have saved me $50.00. Thank you.