Deriv off

The equipment is showing Deriv.off on display and we can not see the signals from patient, could be the cable damaged?...thanks for any help ECG Mortara Eli 250c

   RE: Electrocardiograph (EKG / ECG) Report Post
Fri Mar 05 2021Reply from Ben C
Ben C

Southeastern Biomedical may be able to assist. 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Fri Mar 05 2021Reply from TSS Jim
avatar placemark

I have never seen that message. Broken wires will give a Leads off message. You can take a picture and sent it to Hillrom tech support. mor_tech.support@hillrom.com

Hillrom bought Mortara Instrument a couple years ago.


Report Post
Fri Mar 05 2021Reply from MedRepair Rx, LLC
MedRepair Rx, LLC

I believe that error is a calculation error in the mainboard. Derivative errors could be a processing error or interpret error. Derivative, in mathematics, the rate of change of a function with respect to a variable.... Derivatives are fundamental to the solution of problems in calculations. You will most likely need a mainboard



MedRepair Rx, LLC

Office: 864-757-8799

Fax:    864-757-8798

www.MedRepairRx.com

info@medrepairrx.com
Report Post
Thu Mar 11 2021Reply from medelektron
avatar placemark

!45;09B5 --B5AB, HCELI 250 / ELI 210 @C:>2>4AB2> ?> B5E=8G5A:>1A;C6820=8N.


Report Post
Thu Mar 11 2021Reply from medelektron
avatar placemark

Do an ECG Test, Noise Test identify the faulty channel. ELI 250 / ELI 210 Service Manual.