Deriv off
The equipment is showing Deriv.off on display and we can not see the signals from patient, could be the cable damaged?...thanks for any help ECG Mortara Eli 250c
RE: Electrocardiograph (EKG / ECG) Report Post
Fri Mar 05 2021Reply from Ben C
Southeastern Biomedical may be able to assist.
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
Fri Mar 05 2021Reply from TSS Jim
I have never seen that message. Broken wires will give a Leads off message. You can take a picture and sent it to Hillrom tech support. mor_tech.support@hillrom.com
Hillrom bought Mortara Instrument a couple years ago.
Report Post
Fri Mar 05 2021Reply from MedRepair Rx, LLC
I believe that error is a calculation error in the mainboard. Derivative errors could be a processing error or interpret error. Derivative, in mathematics, the rate of change of a function with respect to a variable.... Derivatives are fundamental to the solution of problems in calculations. You will most likely need a mainboard
MedRepair Rx, LLC
Office: 864-757-8799
Fax: 864-757-8798
www.MedRepairRx.com
info@medrepairrx.com
Report Post
Thu Mar 11 2021Reply from medelektron
Thu Mar 11 2021Reply from medelektron
Do an ECG Test, Noise Test identify the faulty channel. ELI 250 / ELI 210 Service Manual.