Wed Mar 10 2021

Converting 50hz to 60hz

hey, guys . I got this Hologic insight FD from over seas and its running on 50hz. Does anyone know to convert the unit from 50hz into 60hz? Thanks

   RE: Mobile C-Arm Report Post
Thu Mar 11 2021Reply from Amscopolaris
Amscopolaris

no converter required, because switch mode power supply available


Report Post
Thu Mar 11 2021Reply from Medical Tech
avatar placemark

Thank you. Do you know where the switch mode located? i look on the AC input module on the rear cabinet but dont see anything.


Report Post
Fri Mar 12 2021Reply from ricardsc2
avatar placemark

Good morning..you d'not need any switch the power suply is dimensioned to work at 50 or 60 Hz without any change


Report Post
Fri Mar 12 2021Reply from Amscopolaris
Amscopolaris

Suitable for your own power outlet,  the proper power plug is required


Report Post
Fri Mar 12 2021Reply from Medical Tech
avatar placemark

got it. thanks for the help guys


Report Post
Thu Jan 06 2022Reply from Biomedic-device
avatar placemark

Hi i need a service procedure for calibratig the Fluoroscan FD 

 


Report Post