Booting Issue

Hi Everyone, I've been working on this system for a week by now, issue: 1. Breaker ON, 2. Blue light on the AC tray power supply ON, 3. Push the ON button at the front, 4. System boots for 5 sec then immediately shut down, blue light on AC tray blinks. Replaced following: AC tray power supply, platform & Acquisition power supplies and the UMB assembly Issue persists, no progress at all in getting it to boot. Any help or ideas, please let me know.

   RE: Ultrasound Systems Report Post
Thu Mar 11 2021Reply from Bobbybg2
avatar placemark

replace the Acquisition Power Distribution Board on back wall next to umb


Report Post
Thu Mar 11 2021Reply from JAlia
avatar placemark

Sorry, forgot to mention in my initial post, the APDB has been replaced too, the issue remains unsolved. I am just running out of ideas!


Report Post
Thu Mar 11 2021Reply from Bobbybg2
avatar placemark

Unfortunately every time I ever had the issue it was the Acquisition Power Supply, are you a 100% the Acquisition Power Supply you got in and replaced was a good one. I will look at the wiring diagrams and see what I can figure out but if you get it fixed please let us know. Maybe give david a call at aue, he knows those machines really well


Report Post
Thu Mar 11 2021Reply from JAlia
avatar placemark

Well, I tried the APS on a working system and it is good and boots all the way.


Report Post
Thu Mar 11 2021Reply from BMET-KY
avatar placemark

I have had similar issues and its been the power supply voltages being out of spec. I know you mentioned you place them but I would double check your voltages are good coming out of the power supplies. Also check conections I have had connectors get burnt pins causing voltage drops.  If the power supply voltages are good you may be able to check voltages at APDB. You may have bad wiring or connection causing a voltage drop.


Report Post
Thu Mar 11 2021Reply from JAlia
avatar placemark

We actually purchased the whole assembly for the APS, PPS, and APDB and tried it on a working system and it's good and boots normally, but same does not work on the dead system


Report Post
Sat Mar 13 2021Reply from JAlia
avatar placemark

Hello

The system shuts down after 4 seconds, the AC relay in the AC tray power supply does not latch. how is it possible we measure the voltages on the APDB?

Thank you


Report Post
Mon Mar 15 2021Reply from Danyhelp4321
avatar placemark

Check the cable going from the power supply to the AFP (front card cage cover). I've seen those pins burn out and cause issues.


Report Post
Mon Mar 15 2021Reply from JAlia
avatar placemark

Thanks Dany,

 

I actually disconnected all the cables from the power supply assembly to the front end front plane board so to isolate the issue, but still no luck yet. We also replaced the BTM board today and still no progress, very interesting!


Report Post
Wed Oct 02 2024Reply from JGrand
avatar placemark

Was a solution ever found to this? I have an iU22 with the exact same issues


Report Post