Has anyone had an issue with the C-Arm drifting.? I have looked at the service manual but cant find anything.
Thu Mar 11 2021
Hologic Insight
Thu Mar 11 2021Reply from outinleft
Thu Mar 11 2021Reply from Gons MedWrench
The arm will not stay in any position unless you lock it. Is that normal?
sorry i did not mean to post 8 different times. my internet was not working good. Thank you for replying.Report Post
Thu Mar 11 2021Reply from outinleft
1. any particular reason for posting your answer 8 times?
2. if you are talking about it drifting down toward the floor, you need to replace the arm, as the gas tubes are weak. Hologic advises against repairing the arm in the field. We have replacements on an "exchange" basis. What is your serial # ?
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Thu Mar 18 2021Reply from Minicarm.com
yeah you just need a new flex arm for your Hologic InSight 2. Do you have the older style flex arm (metal side covers) or the newer style which has plastic covers on the side?
Joshua Bacon
Vice President
PH: 1 (800) 643-2998
www.minicarm.com
Report Post