Pa out BP22
RE: Modules Report Post
You should be able to, yes. You need to press the PRESS button and define that module as PaP (Pulmonary Artery Pressure), then of course connect it to the correct transducer. Then the output is at the ratio of 1V / 100mmHg, so it will be a very tiny wave form on PA. It cautions that if you select the 6 or 10 mmHg scales, that there is a positive voltage offset of 150 mV that is injected into the output, but I suppose that was to make it more human readable.
And the cable to connect to it is a 3.5mm MONO cable. Tip and ring only. I found one cable that we have that was made by Baxter, and it's 3.5 mono on both ends.
You have to realize that the original intention of the Press Out was to operate a balloon pump, so it was set up as ART. But it should work. Good luck.