Not able to reach READY state during PRE
RE: Analog Radiography Report Post
What is the name of the manufacturer and model of the equipment you are working on?
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
There are a rotor and expose buttons on the Mux board when you open the generator cabinet and look at the high voltage tank. I would see if those work to see if the problem is in the console or the Mux board. Make sure you have one of your console settings is for general radiology so you can bypass the Bucky system. If that works then you got a issue with setup in the console or your not getting a Bucky ready signal from the wall or table buckys.
Report Post
Hi Sir,
Thanks for your prompt responds.
I forget to mention this in my previous report which I tried to make exposure using the switches on the UD CONT board in the Generator cabinet but still getting the same problem. Also I tried using the exposure button the console panel but having same problem. I tried making exposure using different Technique (GR, wall bucky, Table Bucky) but still having same problem.
Please advice.
Regards,
MJack.