Monopolar 1 footswitch not working

What could cause the monopolar 1 (cut)footswitch to stop working

   RE: All Categories Report Post
Thu Mar 18 2021Reply from Amscopolaris
Amscopolaris
What is the manufacturer and model name try the footswitch on another device If the problem persists  switch for the pedal is defective.

Report Post
Thu Sep 05 2024Reply from SEL95
avatar placemark

I have the same question, it's for COVIDIEN Force Fx Electrosurgical Generator. I connect the footswitch and the cut triggers the coag and the coag doesnt trigger anything, it does this on both monopolar 1 and 2. But it works fine with the handswitch. Im doing a pm on the Rigel Uni-therm electro surgical analyzer.


Report Post
Thu Sep 05 2024Reply from SEL95
avatar placemark

So the issue was actually that I was using a Conmed footswitch on a Covidien, the footswitch has to be for the same make they are not interchangeable. Hope this helps anyone.