Start Failure

Hello, I ask for your help with a Lorad m-iv, the equipment turns on and arrives at its startup screen, then when it goes to the screen to admit patients, the equipment re-initializes and remains in this indefinite cycle ... I assume it was corrupted the software... I would be very grateful if you can help me with this case ... I don't have the equipment software Thanks

   RE: Mammography Report Post
Fri Mar 26 2021Reply from Jodiflo
Jodiflo

There are a few possibilities here but none of which I would recommend you try . Please call Hologic and open a service call. It might be possible for them to dial in and fix the issue but it might require a site visit from a FE as well. 
 

Hologic Breast Health Support

877.371.4372


Jodi Andrews - Official MedWrench Guru 
Report Post
Sat Mar 27 2021Reply from quince medical
avatar placemark

Dear friend, this case is more common than you think, in the m4 console (it is not possible for the software to be corrupted) your problem is I think it is due to the power supply that is below in the console, verify that it is sending 5v  (There is a potentiometer that regulates the voltage from the power supply to the card) if you need more information I can gladly help you in detail


Report Post
Tue Mar 30 2021Reply from Jodiflo
Jodiflo

Selenia software can most definitely be corrupted. As a Hologic FE working directly for the manufacturer I assure that can and has happened. You should call Hologic and proceed with them. 


Jodi Andrews - Official MedWrench Guru 
Report Post
Sun Apr 11 2021Reply from am185168
avatar placemark

Hola. Recién puedo responderles. Muchas gracias a todos por su atención.

Jodiflo, no se de donde es ese número telefónico. Soy de Argentina. Conoces alguna forma online de contactarse con ese servicio de Hologic?

 