What yearly maintenance should be done to the revlite SI
Fri Mar 26 2021
Maintenance
Fri Mar 26 2021Reply from Wayne-UML
There are a set of filters inside that should be changed. Changing those out gives you and opportunity to be sure the heat exchanges are free of dust. Verify the debris shield in the hand piece is clean and replace if needed. Finally, you should measure the output of the laser to be sure the energy output is within tolerance and calibrate if needed.
Wayne Delgado
Senior Laser Engineer
www.umlasers.com