I have Siemens - Mobilett XP Hybrid mobile x-ray machine
When I make a shoot the Error 6 and 11 appears .
What I should do , anyone got the service manual , please send it to me
THANKS
Sat Mar 27 2021
Error code 6 + 11
Tue Mar 30 2021Reply from Tomelgy
Error 6 is a filament error.
Error 11 is over voltage at the capacitor bank.
Check to see if you are getting voltage on the capacitor bank, is the capacitor bank charging from power-on?
Che F6 on D927, check F1 on D952.
Report Post
Mon May 27 2024Reply from eadel43_2205
Hello sir
I have a problem with Siemens mobilett xp Eco 2015
It's showing error 6 & 13 & 11
When I changed the fuse on the board D952 only error 11 disappeared
Report Post
Mon May 27 2024Reply from Tomelgy
See above info error 6 is filament problem.
Check for other fuse failures.
This could be indication tube filament is open.
Check filament voltage supply &/or remove collimator to view filament if possible as the tube output port may be covered.