VOLUSON E8 BT18 POWER SUPPLY FAN FAILURE

Hello! VOLUSON E8 BT18 FAN FAILURE After a few hours of hard examination work, the scanner stops and the following message pops up on the main display. The power source reported about accident. PS Code (0x5461) - Data (0100) - FAN1 Defect ----- It happens daily. Only shutdown and reboot help. Has anyone of you met this? Any solution? I suppose that the scanner is overheated. There 3 FANs on the back power supply. Is this 1 of 3 Fans broken? Or a fan controller in the power supply? I will appreciate you advices.

   RE: Ultrasound Systems Report Post
Thu Oct 20 2022Reply from ddkazakov93
avatar placemark

Hello! We have same problem. Do you fix it?


Report Post
Mon Jan 23 2023Reply from Petco
avatar placemark

Hello! Yes. I solved. You will have to replace the main power supply. Completely.