RF out put test

I am having trouble getting appropriate readings while testing RF output. I am using a BC Biomedical ESU-2400 analyzer and according to their tech support I have things connected properly. Olympus says I have the connections right as well but I am getting virtually no output. Any ideas?

   RE: Electrosurgical Unit (ESU) Report Post
Fri Apr 02 2021Reply from Amscopolaris
Amscopolaris
The Cable you are using may be faulty. And the resistance value is also important

Report Post
Tue Apr 06 2021Reply from Periwinkle
Periwinkle

Hi Xaq,

Before calibrating, make sure if you are getting check the RF output first. If there's no RF output then it can be a hardware fault so no need to use an analyser. If there's some RF output: low or high (inappropriate) then only proceed with calibration.

 

Hope this helps. Happy to discuss more.

 

Thanks & Regards,

 

Periwinkle Kaur, Dcads International


Dr Periwinkle Kaur CMgr
Direcor, ESU Medical Solutions Ltd
www.esumedical.co.uk
sales@esumedical.co.uk

Report Post