Fri Apr 02 2021
error 303 and 304 codes simultaneously.
The errors are for FRFS and FLFS for the foot rail switches. I replaced them and they persist until the head is raised above 35 degrees then they magically disappear.. any ideas as to what might be causing this?
   RE: Bed Report Post
Wed Jun 14 2023Reply from FST 2
Notes from my work order that finally corrected the Errors.
Stryker recommended replacing the foot left and right side rail switches.
Replaced foot left and right side rail switches. cleared error log. Errors return when side rails lowered and go away when side rails up.
Stryker recommended replacing the litter signal coil cord.
Replaced litter signal cord. Problem persists.
Stryker who recommended replacing cpu board.
Replaced CPU board. Performed scale calibration, burn in and op test.
Returned to service.