Fri Apr 02 2021
Error 416
Hi Fuji fcr prima T reads 14X14 cassette normally, but it does not read 14X17 cassette and gives error 416, other fcr read the cassette with no problem. I opened the RU and checked the operation with 14X17 cassette it goes smoothly till the ip reaches sc3 sensor the RU reverse move direction and return the ip to the cassette. All cards fuses checked and were ok. Any help appreciated.
Fri Apr 02 2021Reply from Luis_Med
Do you still need help?. You can check the error in the manual.
Regards,
Luis
Tue Apr 06 2021Reply from RTR
It is a process failure. The cassette is still attached to a patient study. Do you know what patient was being done the first time it failed ? If so reopen the study and run the cassette. Or if you remember the patient just start a new study with the same patient demographics and run the cassette. It runs in a different unit because the bar code is not assigned to a study on the other unit.