Tue Apr 06 2021

HAEMATOLOGY ANALYSER

HOW I WANT TO FIX BLANK ERROR?ALREADY CLEAN TRANDUCER,REPLACE ALL REAGENT? AUTO RINSE,ALL MAINTENANCE ARE DONE,STILL BLANK ERROR HAPPEN.

   RE: Hematology Analyzer Report Post
Wed Apr 07 2021Reply from Ben C
Ben C

What is the name of the manufacturer and model of the equipment you are working on? 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Wed Apr 07 2021Reply from nazrilabkktwu
avatar placemark

sysmex XP-100..made in Japan


Report Post
Fri Apr 09 2021Reply from Anonnim
avatar placemark

Which parameter value is out of limit?


Report Post
Fri Apr 09 2021Reply from Anonnim
avatar placemark

Which parameter value is out of limit?


Report Post
Fri Apr 09 2021Reply from nazrilabkktwu
avatar placemark

WBC parameter.


Report Post
Fri Apr 09 2021Reply from Anonnim
avatar placemark

Hi,

Firstly chech SRV if there is any leaking, and clean it. If every thing is Ok, the problem should be at WBC electrode. 

Just dismount WBC chamber and you can see if it is Ok. If it is not, you can solder it or replace with the new one.


Report Post
Wed Apr 14 2021Reply from nazrilabkktwu
avatar placemark

thank you so much sir..


Report Post