Filament offset calibration
RE: DR Systems Report Post
Hi Sagax,
"follow a service manual" it's not an helpfull answer. Anyway Tks for your reply and let's get over.
Resuming my experience
working conditions:
- I've selected the "filament offset calibration" via the generator tool
- I'm working with a scope connected to TP7 on KVP Control Board A2 (AY40-003S) and ground
- a Mas Meter has been connected to A17P1-9 and A17P1-10
OFFSET CALIBRATION FOR SMALL FOCUS IS POSSIBILE; flat curve reached with 75 mA as described in Serv. manual
OFFSET CALIBRATION FOR LARGE FOCUS NOT POSSIBLE; encreasing the offset value via the sliding bar I'm reacing a flat curve with a limited value (max value measured 100 mA and/or 1.3 mAs instead of 300 mA and/or 4.5 mAs).in this condition if the control value is encreased the generator gives "Invalid Filament exposure value"
I even tried to adjust the values via the advanced offset calibration menu with a comparable resoults
Do you have any suggestion?
tks in advanced for your reply