The system passes all the booting sequence but there used to be echo failure to display a proper field of view. It seems I will need to change A333-2 ECPR or A333 SBC. How do I get one. Or what could be a solution to the above description. Thank you.

   RE: Ultrasound Systems Report Post
Mon Apr 12 2021Reply from Jim Carr
Jim Carr

Howdy,

If the LEDs 5 and 7 on the I/O Panel are off, then I believe you are correct that it could be caused by either the ECPR or SBC. Check to see if LEDs on the ECPR are on. DS2 should be on indicating it is communicating with the SBC. DS3 should be on to indicate the ECPR is processing data. DS51, 52, 53 and 54 should be on to indicate the power supply voltages are present.

Good luck,