Troubleshoot error code ERR1

Does the error code "Err1" usually mean that the converter board needs replacing? The service manual says that the main voltage is out of range. But our line voltage is fine. Does anyone have some feedback on this?

   RE: Radiography Report Post
Tue Apr 13 2021Reply from Ben C
Ben C

What is the name of the manufacturer and model of the equipment you are working on? 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Tue Apr 13 2021Reply from Mdemerjian
avatar placemark

Sorry.  The unit is a Gendex Expert DC.  Product #: 110-0205G1. Thanks


Report Post
Tue Apr 13 2021Reply from Mdemerjian
avatar placemark

Hi Ben.  Are you there?


Report Post
Tue Oct 18 2022Reply from MateuszD
avatar placemark

got same problem. Any solutions?

 


Report Post
Thu Dec 01 2022Reply from Downs
avatar placemark

We have found that there is a loose wire in the tube head that is causing the error


Report Post
Tue Jun 27 2023Reply from MateuszD
avatar placemark

Hi Downs, 

Thanks for reply. Did you mean the loose connector ? I will check it tomorrow. 


Report Post
Wed Jul 12 2023Reply from MateuszD
avatar placemark

Not in our case, I have swapped the main power supply board with other Xray - and it seems the PCB is faulty. I have replaced all capacitors and err1 still shows.
I found on eBay that someone is sorting that out for few hundred bucks in US (too expensive shipping)

If anyone has got hints - very welcome


Report Post
Wed Mar 13 2024Reply from MateuszD
avatar placemark

I have sorted out the Err1 problem. If anyone has got a problem with the Power Supply or Logic Board of this Xray -  I am happy to repair. 


Report Post