Tue Apr 13 2021
Troubleshoot error code ERR1
Does the error code "Err1" usually mean that the converter board needs replacing? The service manual says that the main voltage is out of range. But our line voltage is fine. Does anyone have some feedback on this?
RE: Radiography Report Post
Tue Apr 13 2021Reply from Ben C
Tue Apr 13 2021Reply from Mdemerjian
Tue Oct 18 2022Reply from MateuszD
Thu Dec 01 2022Reply from Downs
Tue Jun 27 2023Reply from MateuszD
Hi Downs,
Thanks for reply. Did you mean the loose connector ? I will check it tomorrow.
Wed Jul 12 2023Reply from MateuszD
Not in our case, I have swapped the main power supply board with other Xray - and it seems the PCB is faulty. I have replaced all capacitors and err1 still shows.
I found on eBay that someone is sorting that out for few hundred bucks in US (too expensive shipping)
If anyone has got hints - very welcome
Wed Mar 13 2024Reply from MateuszD
I have sorted out the Err1 problem. If anyone has got a problem with the Power Supply or Logic Board of this Xray - I am happy to repair.