Tue Apr 20 2021

Error code

Showing on tha Error 526

   RE: Pulse Oximeter Report Post
Wed Apr 21 2021Reply from Chris S
Chris S

Error 526 is "SpO2 module not responded".  I would start by replaceing the NELL-3 Spo2 module.


Official MedWrench Guru
Report Post
Wed Apr 21 2021Reply from T-Lobato
T-Lobato

They have a very good service manual which can be located under documents.  It says, 526 is a "SpO2 module not responded".  It is an internal component which can be seen on the diagram on page 73.

T.


TJ Lobato | MedWrench Guru
Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.