Ground?

Can't find a good ground for EST......don't really want to crack this thing open

   RE: Electrosurgical Unit (ESU) Report Post
Thu Apr 29 2021Reply from Chris-CBET
avatar placemark

try using a screwdriver to reach the case screws. 


Report Post
Fri Apr 30 2021Reply from ONizzle
avatar placemark

Unfortunately I tried that, Every metal piece accessible on this unit is un-useable.


Report Post
Fri Apr 30 2021Reply from Cipher
Cipher

The Hyfrecator 2000 uses an isolated power supply and the AC ground (wire) connects only and directly to the power supply board.   There is no other connection point on the case, thus a traditional ground electrical safety check will not work nor is it required.   Conmed doess call for a DC Isolation Test however, which will verify that the patient output jacks are isolated at DC from earth (ground) and from one another.   The test consist of verifying infinite resistance between P/P jacks (and other points) to AC inlet ground on the back of the unit.   The procedure is laid out in the service manaul.   Hope this helps.  - Cipher


Report Post
Fri Apr 30 2021Reply from Chris-CBET
avatar placemark

unplug it, test the resistance of the cord and that's all you can do. 


Report Post
Sun Dec 19 2021Reply from Stan Skaggs
Stan Skaggs

The return ground uses the power ground wire.  You measure from the probe tip to the earth ground.  This is stated in the service manual for doing electrical safety.  Also says should not read higher than 40uA.

Thanks!

mrstan


mrstan.
Report Post
Tue Sep 06 2022Reply from wilmagraves
avatar placemark

It may be possible to access the case screws using a screwdriver.

geometry dash


Report Post