NimaAryamehr

Often this problem solve with  Flow  calibration 

BUt check hpg and Dialis transduser 


Report Post
Mon May 03 2021Reply from reyscorpstech
avatar placemark

How to flow calibrate? Sorry we don't have technician.. thanks a lot


Report Post
Mon May 03 2021Reply from reyscorpstech
avatar placemark

How to flow calibrate? Sorry we don't have technician.. thanks a lot


Report Post
Mon May 03 2021Reply from Deji
avatar placemark

Calibrate hpg transducer and dp transducer.check your flow pump


Report Post
Mon May 03 2021Reply from Deji
avatar placemark

It can also be caused by bad oring. Do variable calibration.check Dialysis flow


Report Post
Mon May 03 2021Reply from reyscorpstech
avatar placemark

You mean Couplings Orings? Thanks a Lot


Report Post
Mon May 03 2021Reply from reyscorpstech
avatar placemark

You mean Couplings Orings? Thanks a Lot


Report Post
Mon May 03 2021Reply from Deji
avatar placemark

Yes. Do DP and HPG transducers calibration. Pls check Deariating chamber  also


Report Post
Wed Aug 18 2021Reply from reyscorpstech
avatar placemark

Please Help

p prec 001 001
P prec 002 000
P prec 001 000